And there is the correct answer..........Miskiet wrote:You get K type and J type thermocouples (probes) amongst a few others. They are used for different temperature ranges/applications. The instrument should have a set-up for changing to either (Most probes we use are K -type) - I would guess yours is set up for J-type and that is why you have got the off-set temps. Alternatively you can contact the manufacturer and find out what the correct type of probe is for your instrument.
Cutting the wires shorter should not make a difference as long as both connections are at the same temperature - Be careful if you cut them (the wires are very fine copper strand - difficult to make good connections Thats why they don't recommend cutting it)